意思是要把itc板洗干净然后交换吗????
|
啊,传感器那玩艺是我理解错了,不好意思各位。可上次我把连接在ITM传感器另一端的栓子拔出来确实提示油墨温度不对,难道是橡皮布和油墨温度同时出现我只看到一个:shame:
还有12楼总版主老大说的制冷机要停机4天以上才可能会出现R134a液化,可我上次日停了一天呀:shocked: 难不成这小玩艺早就想罢工不成。。 嘿嘿,另外再借这求教下下,indigo机器的字库有办法安装自己想要的不,或者说用软件做出来可变数据能转曲不,可变数据的数据库是用Excel做的 上次做个可变的号码机器上没这字体,又安装不进去,只能替换, 太闷了:bad_smile: |
no,you should swop the solenoid valve first.
|
引用:
引用:
一般都是阀的问题。 |
引用:
|
我把整个ink tank 换了,结果发现不只是青色过热,其他颜色都过热
只是青色热得比较快,那怎么办????? |
引用:
你先检查下,看看水温正不正常,油温正不正常再做下一步判断。 |
why don't you swop valve first? i told you it's no relation between your problem and TIC board. the temp sensor on TIC board is more senstive , we can trust it.
as you mention that all inks have this problem ,i am thinking about chiller may have some problem ,you need to check the water pump outlet pressure and water graded filter pressure, the chiller work condition etc. and,you need to verfity that the image oil cooling rate is good or not , it indicate us two things,one is the oil filter,another is chiller work condition .if the oil filter is blocking too much,the image oil can not be cooling down as normal, so the image oil in oil tank will warming up fast when you print something.but even so , it still can not help us to troubleshot your problem , because the image oil can not affert ink temp too much. ok,let's go back to another things about image oil cooling down,the chiller work condition,if the chiller work condition is bad ,you will not only see ink temp problem but also image oil cooling problem .so that's the reason i want you to verfity the image oil cooling rate . as your unknown chiller model , i can't give you further suggestion about chiller but clean condenser. in addtion , the cyan ink's temp is more higher than other inks because the cyan BID is nearest BID to ITM. |
Don’t zhuangbility,zhuangbility leads to leipility!:bad_smile:
|
:grimace:
一说Indigo维修 一大堆总版主啊 潜水艇啊 都浮起来了 插图是要先上传再插图片吧,太麻烦。要不就给个图片你看看了 |
几天都上不了网,急死我啦,我公司是用4色的,现在只有黑色没有过热,其他(红,青,黄)都过日,我应该怎样检查,可以说详细一点吗??因为真的不太懂....
|
chiller water pump pressure
chiller water filter pressure chiller condenser blocking etc |
你可以把默灌里的主板拆下来清洗一下有可能是感应器那里太脏
|
都洗了,不行.........
|
chiller water pump pressure
chiller water filter pressure chiller condenser blocking etc 应该要怎样检查啊???因为我没有碰过那chiller.... pump pressure 再哪?? filter pressuer 再哪??? 还有condenser呢??? 我真的什么都不懂....麻烦啦 |